Transmissive pH Sensing in Microtiter Wells
Introduction Thru the use of a mathematical algortithm the sensor can detect pH changes down to 0.01 pH units. Smart pH Cuvettes are semi-disposable and require very little maintenance. The Sol-Gel BCG has been coated inside microtiter plates for the detection and continuous monitoring of pH changes over the biological range as a proof of concept to be used for the monitoring of cellular pH environments.
Case Study As a quick feasibility study, a brief titration was performed using two buffers only: pH 1 (absorbance reference) and pH 11 (observed). The pH 11 buffer produced a strong, though distorted, absorbance curve very similar to what is seen for the Smart pH Cuvettes, which showed that these wells had the potential to work as hoped:
This was performed using a standard cuvette holder turned vertically, with collimating lenses on each side. Light was fed to the top of the cuvette holder from an LS-1 light source with a blue filter via a P200 UV/VIS fiber. The signal was then sent through a P300 UV/SR fiber to a USB2000 with grating #2, 200um slit, and no lens installed. The experimental setup is shown below:
However, when our customer carried out their own set of titrations, the data did not adhere to the pH algorithm whatsoever; everything appeared very randomized:
As a result, a full titration was performed here at Ocean Optics to determine the cause of this significant error. The same fibers, light source, spectrometer, and general setup were used as previously described. Two issues were determined to be the most probable causes of the poor data seen by our customer; these included the lack of collimating lenses being used at the facility, as well as an incorrect baseline correction wavelength. They were previously instructed to use 750nm as the baseline correction found in the algorithm, though they used 880nm due to availability of that filter. The tests performed here showed that neither 750nm or 880nm were acceptable baseline wavelengths, rather the ideal wavelength to use was 509nm. Results from the full titration are shown below:
This has shown that the pH sensor affixed to the bottom of microtiter wells can accurately monitor and report the pH of transmissive analyte solutions.
Algorithm used:
|
FL-400